[Techtoolslist] Fluke 9010A Input voltage
Filip Hautekeete
fhaut at griffion.com
Mon Oct 31 16:46:39 EST 2005
Hello Rodger,
You are correct I should replace them asap. Thanks for pointing this out.
Best regards,
Filip
----- Original Message -----
From: "Rodger Boots" <rlboots at cedar-rapids.net>
To: "Technical Tools Mail List" <techtoolslist at flippers.com>
Sent: Monday, October 31, 2005 10:40 PM
Subject: Re: [Techtoolslist] Fluke 9010A Input voltage
> Be careful, a lot of times when you change voltage you also should
> change the fuse.
> In the case of the 9010A the line fuse should be 1A fast blow for 120 V
> and 1/2 A fast blow for 240 V.
>
>
>
> Filip Hautekeete wrote:
>
>> Hello Steve,
>>
>> You are correct. Thanks to John Robertson I figured out there is a
>> slide card for voltage selection behind the fuse panel. No need to
>> drag the systems out of your garage ;-)
>>
>> Thanks for your help!
>>
>> Filip
>
>